Hi John, dear NI support:
a NI application engineer wrote:
"The response time of a thermocouple is really zero because it's a voltage source"
oopsy
John, take care. This is not correct. Certainly every T/C has a response time. Not an electrical one, but a thermal one. Imagine you have your T/C connected to a heat plate. When the heat plate changes the temperature, the T/C will need some time to reach the same temperature. The time constant for that process is proportional to the mass m of the T/C and the specific heat capacity C, but inverse proportional to the heat conductivity K.
The time constant is usually larger for gas measurements than for liquid or solid contact measurements. The time constant also critically depends on the thermal contact between the T/C and the object of interest.
Once more: Each thermocouple has a certain response time. which depends on a lot of parameters. Certainly, the smaller the T/C the faster the response time. You want to measure temperatures of your object with 2kS/s (500 us). You are looking for a real tiny T/C.
Of course, the T/C voltage can be sampled even with 333kS/s and the reading will reflect the actual (average) temperature of the T/C junction - but this is not the temperature of the object of interest. You are looking for a tiny, fast T/C (demanding). On the DAQ side you evenmight use your PC's microphone input to sample 16 bit, 2kS/s (not so demanding).
Nepu.