Sampling rate per analog input on multichannel scanning

Solved!
Go to solution

Hi,

 

I expect to buy a DAQ system soon. I have the following spec for the kind of DAC that I'm looking for:

- At least 50 channels (analog inputs with common reference could work),

- Continuous scan with sampling rate of 24kHz or higher.

 

The kind of measurements that I have to perform is compatible with your USB-6255 device.

But I don't get from the specification of this product if the sampling rate per channel is a direct function of the fastest sampling rate available (1.25 MS/s for 1 channel) divided by the number of channels during the coontinuous scan. If so, this means the sampling rate per channel would be 25kS/s for 50 channels.

 

Can you provide a clear statement about it?

My expectation for a clear statement is the following:

I want a way to calculate by my own the sampling rate per each channel to understand if the device could be a good purchase.

 

Regards.

 

0 Kudos
Message 1 of 4
(2,625 Views)
Solution
Accepted by crossrulz

Hello pat.langlois,

 

page 1 in the specification of the NI 6255

Multichannel maximum (aggregate) = 750 kS/s

This means in your case 750KS/s / 50 = 15 KS/s

best regards
Alexander
Message 2 of 4
(2,569 Views)

Hello pat.langlois,

 

if you want usb you can use 5x NI 9205 in a cDAQ 9178.

best regards
Alexander
0 Kudos
Message 3 of 4
(2,567 Views)

Thank you Alexander for your input. That's exactly the info that I was looking for.

I consider this case as closed with your answer.

 

Best regards,

Patrick

 

0 Kudos