Read Multiple 9181 Chassis Simultaneously

I am designing a new project and I need to know if I can get away with purchasing 2 9181 chassis instead of one 9188 chassis.  I will be using a 9213 for thermocouple readings and a 9208 for 4-20mA current readings.  The signals dont nee to be syncronized and will only have to read once a minute or so.  My question is whether or not I can read the signals from both chassis at the same time through MAX and easily imput into a LabVIEW program.  

 

I am currently using a single 9188 to accomplish the same tasks at a different location but am trying to cut costs on the next project.  The 9184 is not an option due to the software version I am running. 

0 Kudos
Message 1 of 2
(3,968 Views)

Hi rifenbug,

 

Yes, you can do this easily in MAX or LabVIEW using DAQmx. You could literally just copy/paste togeter a thermocouple shipping example and a current input shipping example, although you'll probably want to customize and optimize a little bit from there. Just keep in mind that by cutting costs now, you're increasing the cost when you want to expand your application to include more than 2 modules. 

 