I want to protect X series (6320) analog inputs from overvoltage. The inputs have a maximum impedance of 100k (usually lower). It can happen that one or two of the inputs are connected to 70 volts. This can happen with the computer on or off. I want to protect the DAQ card from overvoltage damage. It doesn’t matter what the card returns when the voltage is out of range.
I don’t want to load the input when the voltage is between 0 and 10 volts.
http://www.ni.com/datasheet/pdf/en/ds-152
From this document, the specifications are:
Maximum working voltage for analog inputs (signal + common mode) ±11 V of AI GND
Input impedance Device on AI+ to AI GND >10 GΩ in parallel with 100 pF
Input impedance Device off AI+ to AI GND 1200 Ω
Overvoltage protection (AI <0..31>, AI SENSE, AI SENSE 2) Device on ±25 V for up to two AI pins
Overvoltage protection (AI <0..31>, AI SENSE, AI SENSE 2) Device off ±15 V for up to two AI pins
Input current during overvoltage condition ±20 mA max/AI pin ±20 mA max/AI pin
How can I protect the DAQ card? A 10K resistor in series with the AI input, would limit current to 7 ma. This should handle the device off situation. 10K should be insignificant compared to the 10 GΩ input impedence of the DAQ. A Zener diode could clamp the voltage but I am concerned about the leakage current.