10-01-2015 06:59 AM
Hello, i made a program to output 30 waveforms, but when i try to output more then 16 signals, the program show the error Internal Software Error occurred in MIO software. Please contact National Instruments Support. I am using two NI 9264.
Please help me
Thanks for support
Solved! Go to Solution.
10-01-2015 01:06 PM
What cDAQ chassis are you using? The cDAQ-9172 only supports 16 hardware timed channels but the other cDAQ chassis don't have a limit unless you are doing on-board regeneration which you aren't.
What version of DAQmx do you have installed?
You should never get that Internal Software Error so that's an issue on the DAQmx side. I don't have two 9264s at my desk at the momement but I can try it out later to see where the error is coming from.
10-01-2015 01:38 PM
Hello MarkGrot, thaks for your response but i am using a cDAQ-9178, and my DAQmx is the latest version.
10-02-2015 06:38 AM
I read de manual and my Chassis only can write 16 channels thanks for suport!
10-02-2015 11:09 AM
No, your chassis should support >16 hardware timed AO channels. It looks like we broke that feature in DAQmx 15.1. I haven't tested it but it looks like it should work in DAQmx 15.0.
08-03-2016 04:44 PM
As an update, this issue (reported with CAR ID 549088) was fixed in NI-DAQmx 15.5.