NI 9205 card is not recognized by MAX/ "Error 201003 occured at DAQmx"

Hello 

 

I am using MAX version 15 for labview 2015. The program is used for controlling fans, spark ignition and reading temperature and pressure. 5 NI cards such as NI 9205 are used in the DAQ system. After doing an experiment, the labview gave me an error " 201003 occured at DAQmx create channel (AO. voltage. basic).vi ". 

 

For trouble shooting, I openned MAX and found that card NI 9205 has a small x and turned to be not recognized. I checked from Devices manager-DAQ and it showed me that all other cards are detected except Ni 9205. I tried to restart and turn the pc on and off, reseat the card, seat it in another slot but still it wasn't recognized. I seated a working card on the same slot of NI 9205, the working card was recognized.When I do a test for the other cards it gives OK, while for Ni 9205 thee options are in grey colour and test is not available. It seems that whatever I do, NI 9205 wont be recognized. Is it damaged? I hope not.

 

Thanks a lot for any help

0 Kudos
Message 1 of 2
(2,403 Views)

Looks like you've already done a few of these steps, but here is a flowchart you can use to continue troubleshooting.

 

Why Can't I Find My DAQ Device in Measurement & Automation Explorer (MAX)?

http://digital.ni.com/public.nsf/allkb/34CC88F4838CD37E86256A290066A2BE?OpenDocument

Maggie M.
Application Engineer
National Instruments.
http://www.ni.com/support
0 Kudos