Hi,
I have a couple of questions for you.
1) How does the cDAQ show up in Windows Device Manager? (A screenshot would be great)
2) Can you also post a screenshot of NI MAX with Devices and Interfaces expanded?
3) Can you try to connect the cDAQ to a different computer?
4) What Operating System are you using?
5) Which version of NI MAX do you have installed?
6) Do you have a different cDAQ chassis with the same modules that is visible in NI MAX?
7) You mentioned that you have restarted the NI Config service and NI Device Loader, what was their status when you restarted them?
😎 Where did you install DAQmx? From the NI website, or from a DVD?
9) Are other DAQ devices properly recognized by this same system?
Reinstalling LabVIEW should not have any effect on how the cDAQ is recognized in NI MAX.
Below are some additional links that may be helpful to you, not sure if you have seen them yet.
http://digital.ni.com/public.nsf/allkb/2E02F8EC5D0197928625758C005FFD8D?OpenDocument
http://digital.ni.com/public.nsf/allkb/2E02F8EC5D0197928625758C005FFD8D?OpenDocument
Kevin S
Applications Engineer
National Instruments