12-04-2015 04:29 PM
I purchased an NI 9217 and Pt Thin Film RTDs to make a temperature measurement. But I now realize that the RTD is acting as a non-negligible heat flow path in my system and so I want to switch to a thin wire Thermocouple that touches the surface.
But purchasing another module is a huge investment. So I am trying to find what my options are.
Thanks for the help!
12-05-2015 04:02 AM
Read user manual of NI 9217 for connecting RTDs at analog inputs.
Jamal
12-07-2015 03:16 AM - edited 12-07-2015 03:17 AM
Well, the 9217 is made for 100Ohm RTDs.....
If heat transfer is a problem:
Use
-thinner wires
-smaller RTDs
But keep in mind the 1mA excitation(/heating 😉 ) current, maybe that self heating current is too high? (OK, the spec doesn't tell if it is applied all the time.....)
A hack (never tried) with a lot of calibration needs: (and you didn't tell us a lot about your application.... not even the temperature range)
And it can only work if the 9217 measure in DC mode..... (what I assume)
Get 1 to 3 good low tempco 100R resistors, put them together a 100RTD (CH4 😉 ) on a heat shielded metall block to build your CJC . (Everything with good thermal contact)
Wire the 100R Resistors in 4 wire mode, connect your TC in series to the RTD+ connection (these connections are also placed on the CJC block)
Any TC EMF will show up as a change in the measured resistance. For a Pt100 (@ 1mA) 1K temperature change is about 380µV voltage change .... a type K is about 10 times smaller.... uncertaincy is about 1K anyway with a TC, so expect 2K and a resolution of 0.1K ....
How to calc the temperature is now your homework 😉
12-07-2015 03:28 AM
Last remark:
You will spend a lot of time with that hack ( building the CJC block, programming, validation/calibration) ....
so getting a TC module if you need to measure with TCs is the recommended way to go....
03-27-2017 08:29 AM - edited 03-27-2017 08:30 AM
Thank you for the Kudos 🙂
Just curious: How did you solved the problem?
03-27-2017 08:31 AM
I wasn't able to. I eventually had to throw money at the problem and buy a module to use thermocouples