for the community: we solved the problem checking the configuration under
MAX (Measurement and Automation Explorer). We had to set the accessory for
the SCXI-1127 module to "SCXI-1332" instead of "SCXI-1331" (default value):
in this way, we tell the system that the module will be used in Matrix mode.
Hope this will help other users,
Maurizio
"Alex Robles" wrote:
>>Maurizio,>>Thanks a lot for your help, I deleted the chassis and re-installed
it in>the way you told me. The only problem is, I can't manually detect modules,>when
expanding the SCXI 2000 device, I get no listing of slots, this is,>when
clicking on the '+', the list of modules in the chassis is not displayed.>>I
proceeded to auto-detect modules, and the SCXI 1127 gets detected in the>correct
slot, but again when trying to get the listing of modules, it won't>be displayed.>>I
refuse to think the chassis is damaged since it's brand new, what else>could
be wrong?>>>"Maurizio Basso" wrote:>>>Hi Alex,>>sorry
that the mistery is still unsolved: let's see if we can>progress.>>Hmm, I
agree with you that at this point it is useful to verify>that the>module
is actually working (I assume you do not have another module>to try,>to verify
this theory).>>Let's start from the beginning, to make>sure we do not miss
anything.>First, let's verify that the module in Max> is correctly recognized.
I am>a little confused because in the first message>I read that to make the
chassis>working you have to test the chassis in MAX>and then it works, while
in the>second message I read that MAX cannot interact>correctly with the
module>(maybe I am not giving the right interpretation>to your sentences).
I would>delete your chassis from MAX and re-install it,>with right-click
on "devices>and interfaces" -> insert -> Scxi 2000.>Your>are actually advised
to place the module in slot 4 as general guidance:>in>the specific case of
serial control of SCXI chassis (SCXI 2000) any slot>would>be ok.>At this
point, just to try the most manual way, configure the module>SCXI-1127>without
auto-detection, manually specifying that in slot 4 you>have the SCXI-1127.>at
this point, if a right click on the SCXI chassis in>MAX does not allow>you
to have a positive test of the chassis, we might have>a problem with>the
specific module.>>Please let me know.>Thank you,>>Maurizio>>"Alex>Robles"
wrote:>>>Hey Maurizio, thanks for the reply,>but
the mystery is still unsolved.>In>regard to:>>1) In this case, I am actually>using
ni switch soft front>panel, and I've>tried>SCXI1::4::SCANNER>SCXI1::4::MATRIX>SCXI1::4::INDEP(my>SCXI>1127
is in slot 4)>None of them works, basically when I go to reset>or self-test>I
get the message>"instrument session is invalid" .>I am not>in Labview for>this
case.>>2)I get the same message when I test the chassis>in MAX, I've>gone
thru the>Labview examples, none of them works. I am thinking>it could>be
the 1127>module, by the way, shouldn't it be in slot 4 for it>to work>properly?
I>see that you placed it in slot 2.>>3)In Labview, I am>using the>correct
channel names according to the initialization,>and my program>runs>without
errors(except on matrix mode), but it just indicates>that the>path>is not
supported.>>By the way, I can initialize(from Labview) only INSTR,>SCANNER>and
INDEP>modes, but I can't initialize MATRIX mode, I get the error>"id>query
function>failed". Also, when I try to get the firmware version,>I don't>get
a return>string, only the driver version string. Could it be>the module?>>HELP!>>"Maurizio>Basso"
wrote:>>>>Alex>Robles wrote in message <38ff2897@newsgroups.ni.com>...>>>>Hi>Alex,>>>I>recently>got
a SCXI 2000 chassis and a SCXI 1127 Multiplexer/Matrix>module,>>which>I>want
to control through the RS-232 port of my PC using Labview>5.1.>>>>Please>anyone>who
might know how to troubleshoot the following help>me,>I'm>>going>nuts>trying
to figure it out:>>>>1) I can't control the SCXI>1127 module>using>NI-Switch,
I receive an error>>message indicating that>the instrument>session>is invalid
when I try to>>reset or self-test the instrument,>therefore>I>am not sure
if the switching>>functions are operational. what's>wrong?>>>mmhhh,>it>is
not clear if you are creating an instrument handler>before calling>niSwitch_Reset:>this>is>something
that it's worth checking.>ni_Switch_initialize is the function>to>use>in
this case.>>>2) The way I'm>trying to control the multiplexer with>Labview>is
I power>>up>>the chassis>and then open my program, but as soon>as the>'initialize'
vi>>kicks in the>program jams. All the time the 'online'>led>in the chassis>is>>blinking.>>In>order
for it to work I have to run MAX,>self>test the chassis and then>>run>the
'initialize' vi, this way it'll work>and>the 'online' led stops>>blinking.>What's
wrong?>>>*** The fact that in>MAX>you are able to test the chassis,>indicates
that>the configuration you>set>is correct.>I have an scxi-1127>in slot2 of
an SCXI-2000 remote chassis>in>my setup.>My scxi-1127 is configured>for 2-wire
mode and with scxi-1331>accessory.>I>am able to test the chassis>in MAX,
the message after test is>"The chassis>has>been verified">I would>suggest
also to have a look on the>examples that are>on the root>of the NI-Switch>CD
under the directory "examples".>Also,>on>www.ni.com/support, searching>for
"1127", the document "KnowledgeBase>->1TCG91PH>- Using the NI 4060 Digital>Multimeter
with the SCXI-1127 - ...">could>give>some hints as well. >'Hope>this helps.>>>3)
I am initializing the chassis>as>SCXI1::4::INSTR, but when>I use the>>can_connect_channels
vi, no combination>of>channels is a supported>path.>>What am I doing wrong?>>>***
niSwitch_Connect_Channels>will>execute>correctly only if the channel>names
used>are correct with regard>to>the mode>you choose in MAX under the Channel
tab>(under properties of>the>SCXI-1127).>>If you are in 2-wire mode >for
example you can connect ch2>to>com0, or ch7>to com1.>Maybe you are configured
in matrix mode with SCXI-1332>terminal>mode>? In>this case your connections>should
be rows and columns,>something>like>r0 with c3 or similar.>>>4) The user
manual and hardware installation>instructions>do>not indicate>>that the use
of a SCXI-1200 module is needed,>but the NI>SCXI>Troubleshootin>>and Installation
Wizard mentions the SCXI>1200 as one>of>the minimum requirements>>to setting
up a SCXI-2000 chassis.>Which one>should>I go for?>>>SCXI-1200 is only needed
if you are controlling>analog>modules>inside the>scxi-2000>remote chassis.
If your scxi-2000 chassis>is>populated>with digital modules>ONLY (like the
1127, 1160, 1161 etc)>then>you>do not>need the scxi-1200 controller: the
serial connection will>be sufficient.>>>Regards,>>Maurizio>>>Hope>you>can
help me.>>>>-Alex.