9207 current measurement inaccuracy / offset

Hello

 

I am using cRIO-9074 with NI9207 for current measurement from pressure transducer 4~20 mA signal. I have observed that the readings mentioned by equipment are slightly offset when compared to a standard digital multi meter for measuring current.

 

 

 

The pressure transducer was supposed to read 0bar and signal = 4mA, When measured with digital multimeter the reading is 4mA and when i use 9207 module to read the same channel i get 0.0037 which is equal to 3.7mA.

 

Is this the problem with calliberation of i am doing something wrong. All equipment was purchased brand new some months back.

 

Connection diagram is attached below.

 

Please help

 

Best Regards  

Muhammad Usman

0 Kudos
Message 1 of 3
(3,882 Views)

Have you measured in High speed mode or high resolution mode?

 

 

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 2 of 3
(3,870 Views)

I have checked in both both resolution and high speed mode. Currently it is in high resolution mode. I am also considering to check the caliberation of the external digital mulimeter. 