Motion Control and Motor Drives
Softmotion error 3120
I am interfacing an NI-9501 direct-drive stepper module (within a cRIO-9073) chassis. I hacked the example project found in the getting started guide to try and match my system (attached). I got through the FPGA compile okay, and when I go to the interactive test panel for my 9501 axis, it returns the softmotion error 3120 (undervoltage), and the led status light on the 9501 goes to red. I got out the Fluke and verified that all supply voltages are set at 24VDC, IAW with the operating instructions. I'm stumped - I suspect there is something wrong with the FPGA VI that's monitoring the module, but am not savvy enough on FPGA coding to know where to begin. Any thoughts?
(5,566 Views)
Hi there,
The most common cause of this fault is the voltage supply connected to the 9501 itself. Can you confirm that a compatible power supply is connected to pins 7 (COM) and 8 (Vsup) of the 9501?
Cheers,
Andy C.
(5,521 Views)
yes - the Fluke reads 24V supplied to the 9501, IAW with the OI manual
(5,519 Views)
ok - I just checked again.....and the neutral was connected to Pin9, not the COM on Pin 7......looks like it pays to read the wiring diagram!
(5,516 Views)
