Measurement Studio for VC++
Register-based programming on VXI devices s.a. HP E1472A
The problem I had is bit 0 of the status register is stuck at 1 (reset), and according the user's manual, this bit has to be set to 0 to be able to operate the switches. I wrote to this register to set this bit 0 to 0, but when I read it back, it's still 1. I can set/reset other bits of this register but not bit 0. I don't know why. Your help would be profoundly appreciated.
(3,205 Views)
This forum is for discussion of Measurement Studio for Visual C++. You would be much more likely to find someone who knows this in the VXI/VME forum here.
Best Regards,
Chris Matthews
(3,205 Views)
Phuongtran,
Bit 0 of the Status/Control register is defined to be "Device Dependent" for reading and "Reset" for writing. Therefore, if you are believe Bit 0 of the Status/Control register of the instrument is not being properly set by the device after initialization, then it might be best to contact the instrument's manufacturer.
Greg Caesar
(3,205 Views)
Hi Greg,
Thank you for your answer. I am new to Register-based programming so I relied on the user's manual. The user manual says the status register is a read/write resgister; it is at offset 4; to reset the RF Mutiplexer (HP E1472A), write a "1" to bit 0 of the Status/Control Register, but it is necessary to write a "0" to bit 0 after the reset has been performed before any other commands can be programmed and executed.
So I should be able to change bit 0, right?
This device works on VAX system. We are trying to rehost the test program to a PC with a VXI-MXI-2.
Thanks.
(3,205 Views)
