Measurement Studio for VB6
VB6 NI-USB-9237 DAQmxCreateAIVoltageChan selected physical channel does not support the measurement type required
I have a VB6 program with code that properly reads analog inputs from an NI-USB-6008.
I tried to re-use the code for a strain-gage module NI-USB-9237, but at the DAQmxCreateAIVoltageChan step I get the following error:
"Measurements: Selected physical channel does not support the measurement type required by the virtual channel you are creating.
Create a channel of a measurement type that is supported by the physical channel, or select a physical channel that supports the measurement type."
Should I be calling a function other than DAQmxCreateAIVoltageChan?
If so, what is it, or where can I find the reference for such functions?
Or - if that is the right function, should I pass different arguments?  Currently I'm calling:
DAQmxErrChk DAQmxCreateAIVoltageChan(taskHandle, "Dev1/ai1", "", DAQmx_Val_Cfg_Default, -10, 10, DAQmx_Val_VoltageUnits1_Volts, "")
Thanks in advance for your help
-Chris Hardy
(7,810 Views)
Accepted by topic author CHardy
I figured it out from 372251a.PDF
First, the correct function is DAQmxCreateAIVoltageChanWithExcit
Second, you can't just ask for one sample on one channel - ask for 2 samples instead.
-Chris Hardy
(7,808 Views)
