NI USB-9162 Task Configuration Error

Hi!

 

I am tring to read the resistance (in the 2 wires configuration) of a Pt100 sensor. The task is created using Measurement and Automation and apparently working fine. However the Pt100 sensor has a reccomended measurement current of 1.0 mA. While trying to change the task's configuration (internal Iex source) to 1m and trying to run/save the task I am getting DAQ-Assistent Error 200077, value not supported. As it seems, the Iex current is limited only to 500uA. Is there anyway to safely (for the 9162) run an Iex of 1mA?

0 Kudos
Message 1 of 5
(6,245 Views)

Hello o_ptr,

 

NI USB 9162 is a Single Slot USB Carrier for C Series Modules. Since the Carrier can be used for a variety of Modules it would be good to know witch C Series Module you are using?

 

Regards!

 

Moritz M.

0 Kudos
Message 2 of 5
(6,229 Views)

True. Sorry for not being precise enough.

 

The Hi-Speed USB Carrier NI USB-9162 is running with the NI 9219 module (4-Ch Universal Analog Input).

 

Regards,

P

0 Kudos
Message 3 of 5
(6,226 Views)

Hi o_ptr,

 

I tried to find out what settings are possible for this resistance 2-wire measurement. There are no remarks for this in the specification of the 9219. So I would say that there

is no way to change the Iex for this measurement. In this case, the module is a constant current source which only can source 0,5mA. This is always the expected and only supported value.

 

Kind Regards,

 

Salle

0 Kudos
Message 4 of 5
(6,209 Views)

Too bad. Thank you!

0 Kudos
Message 5 of 5
(6,204 Views)