Here are some suggestion for the
EBUS (14) Error.
If the driver and interface are too fast for the intrument you can increase the T1 handshake timing for the interface in the Measurement & Automation Explorer (MAX). By default T1 is set to 500ns. You can increase it to 2us. Right-click on the interface and go to the Advanced tab. use the "Bus Timing" drop list to change the value. T1 is the DIO lines settling time used by the interface when transmitting commands.
This setting can also be configured using ibconfig (with a board descriptor) and the attribute value IbcTIMING (0x0011).
I looked through th
e instrument driver and the code is very simple. You could try diferent things, for example, renoving the ibclr call and see how that affects the operation of the instrument.
Use the interactive control to experiment different functions. Note: with the interactive control you cannot use OpenDev, Use ibdev.
DiegoF
National instruments