reading voltage converting from current in daq

Solved!
Go to solution

Hi all,

i have a pressure sensor that gives ma as output.my daq is reading voltage.i connect resistance as parallel to analog input.But i still cannot read voltage.when i check with ammeter the current value is correct but it cannot convert this current the voltage.

 

0 Kudos
Message 1 of 5
(3,370 Views)

What DAQ are you using?  Are you applying a scale?


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 2 of 5
(3,352 Views)

I am using NI 6229 BNC DAQ. I think i have a problem with connection.Sensor has 2 out.One is for power supply other is for signal.I connect sensor + out to power supply + and signal to ai0 from DAQ.I have to connect power supply - to analog input - but i dont know/cannot find where the ai- in this DAQ

0 Kudos
Message 3 of 5
(3,306 Views)
Solution
Accepted by topic author sgirgin

BNC connections use the outer contact to carry the AI(-).  So, AI(-) should be the round metal cylinder on the DAQ device.

0 Kudos
Message 4 of 5
(3,287 Views)

It works thank you so much i have been working to solve it several weeks 🙂