calibration MMA7361L Accelerometer

Hello,

 

Does anyone now how to calibrate a MMA7361L accelerometer on labview?

 

Any help would be much appreciated.

 

Kind regards,,

 

Aadil

0 Kudos
Message 1 of 3
(2,801 Views)

Static: turn it in each 6 positions , measure the output voltages , do some math ...  look up your local  g  (some say one turn per axis (=3positions) are enough ....)

Dynamicly: place it on a shaker together with a reference , excite the sensor, measure .....

 

there is at least one appnote from AD or Freescale on earth gravity cal. ....

 

If you do this ON a LabVIEW DVD one migth say that you have done it on LabVIEW 😉  

Oh it's friday 😄

 

Greetings from Germany
Henrik

LV since v3.1

“ground” is a convenient fantasy

'˙˙˙˙uıɐƃɐ lɐıp puɐ °06 ǝuoɥd ɹnoʎ uɹnʇ ǝsɐǝld 'ʎɹɐuıƃɐɯı sı pǝlɐıp ǝʌɐɥ noʎ ɹǝqɯnu ǝɥʇ'


0 Kudos
Message 2 of 3
(2,784 Views)

Hello Aadil,

 

I'm not very familiar with the LabView aspect, but can help with the measurement technique.

 

For a static measurement, turning it in the 6 positions as suggested by Henrik is a good method. Gravity is a physical constant (with some variation based on location) and will give you traceability directly to the SI.

 

When performing a static measurement, we generally like to use gravity to apply a +1 g and -1 g to the axis being tested (sensitivity is the difference in output at these two positions divided by 2). The 0 g orientation is generally avoided unless there is some practical limitation requiring us to use it (like the cable/connector being in the way).

 

Why? Because the true input applied to the sensing element is far less sensitive to errors in the orientation of the sensor in the +1 and -1 g orientations compared with the 0 g direction (cos(0 +/- x) vs. sin(0 +/- x); where x is the position error; for small values of x, the difference in the "cos" value is close to 0 whereas the difference in the "sin" value will be almost equal to x).

 

If you are using the sensor dynamically, calibration gets a little more complicated. ISO 16063 Part 21 covers the method, but you will essentially be learning dynamic calibration, shaker control (which can damage the shaker if done improperly), and a far more complicated uncertainty analysis (at least if you want to do it right). Narrow-band measurement techniques such as DFT or FFT are also recommended.

 

If you do not already have this knowledge base, you may want to consider contacting a local metrology lab (if you have a small number of units) or consider a commercial calibration system (if you have a larger number of units). We do both of these and are happy to work with you if you decide that dynamic calibration is needed.

 