This error occurs in NI MAX, version 16, and LabVIEW 2015.
I have four c-modules plugged into the cDAQ-9171, all of the modules are analog voltage in, for example, the 9223, 9234, 9239, and 9244 modules.
Here are the steps to recreate the error:
It does not matter what order I click on the modules, whenever I get to the fourth one I get this error.
So does the cDAQ-9171 only allow 3 analog input modules? I really do not understand this error. I have tried unplugging modules, using different modules, etc. What am I missing?
Cheers,
mcduff
Solved! Go to Solution.
03-06-2017 03:51 PM
Yes, according to datasheet you have 3 timing engines = 3 tasks with different frequency.
You can join 2 tasks for 2 modules into 1 task and sample at the same frequency.
03-06-2017 03:54 PM - edited 03-06-2017 04:01 PM
Thanks for clearing that up. One last question, where on the data sheet does it state this?
cheers,
mcduff
EDIT: I found it, I was looking in the manual before, not the data sheet.
Thanks again.
03-06-2017 04:03 PM - edited 03-06-2017 04:03 PM
It is not in 9174 specifications (http://www.ni.com/pdf/manuals/374045a.pdf)!
It is in 917x overview, page 1:
http://www.ni.com/datasheet/pdf/en/ds-178
and in 9174 manual, page 27 Analog input