07-26-2012 07:03 AM
I am using a NI USB 6211 DAQ with LabVIEW. I am trying to send an output through one of my AO pins, which is connected to one of my AI pins. I want to then read in the voltage from the AI pin. However, anytime I have something wired to my AI pin, regardless of whether it is powered, it registers a voltage of 10.91 volts. Any ideas?
Solved! Go to Solution.
07-26-2012 07:54 AM
Do you have the device set up properly?
Have you run self-tests through MAX on the device?
What do you see when you access the test panels through MAX?
07-26-2012 08:20 AM
Runnin a panel test in MAX yeilds the same results
07-26-2012 08:29 AM
Call NI tech support and determine if the device is damaged; sounds like it may be.
07-26-2012 08:38 AM
Hi
what was the result when you ran it for the FIRST time?
Did you try 'Reset Device' in MAX or by the DAQmx Reset Device function?
07-26-2012 08:42 AM
Freelance,
I have tried resetting the device. As soon as I disconnect the wire while running my measurement vi, the voltage at the input pin begins dropping back down to 0V, but instantaneously, when I touch a wire (can be connected to ground, power, or another terminal on the device) the voltage jumps back to 10.91V
07-26-2012 08:53 AM
Ok
have you tried changing the input channel that is configured?
Did you try connecting a known power supply to this input channel, to check if the input is read correctly?
say, connect a 5V dc input of a frequency generator output of 1Hz sine wave?
07-26-2012 09:23 AM
Yes, I have tried several other channels. I have tried supplying a known DC level of 5V and have tried just shorting to ground. Even if I just touch the wire with my finger, the read voltage suddenly jumps to 10.91V.
07-26-2012 09:31 AM
Are the grounds properly tied together? Meaning AI Sense should be tied to the ground of your signal.
07-26-2012 09:39 AM
Crossrulz,
Just to make sure I understand - you are saying tie AI Sense, AI GND, AO GND, DGND and my power supply ground together?
Won't that create a ground loop?