Why do I get error 10080 with SCXI Cal Constants?

I'm trying to calibrate a scxi 1126 with a 1327 terminal block, attached to a 6024E PCI board.  7 of the 8 channels are within spec, but one channel is slightly out.  I followed the calibration procedures, but when I went to use the SCXICalConstants.vi, I got the error:
 

Error -10080 occurred at SCXI Cal Constant

Possible reason(s):

NI-DAQ LV:  The gain or gain adjust is invalid.

 

How do I know which gain is invalid (DAQ, SCXI, TB), and how do I figure out the correct value?  I've tried many combinations of what I thought would be valid gains, and I've tried to just read the current calibration constants, all with no success.

 

0 Kudos
Message 1 of 2
(2,353 Views)
I figured it out.  Apparently the SCXI gain control needed to be set at a value of 0.02.  This corresponds to the frequency range of 250 Hz.  Although I'm not sure why this value works where 0.00125, corresponding to 4KHz did not. 