06-24-2015 01:42 AM
Hello everyone,
I'm using NI PXIe 5665 (3.6GHz) RFSA which uses 5603 downconverter, 5653 synthesizer and 5622 IF digitizer. I used the NI-RFSA Soft Front Panel initially to acquire the RF spectrum with Reference level: 0dBm, Center Frequency: 100MHz, Span 50MHz and RBW: 3kHz. There was no input given to the downconverter RF In port. The received spectrum is shown below which is the expected spectrum because there was no signal present at the input of the downconverter.
Then I opened up the example 'RFSA Getting Started Spectrum.vi' from the NI example finder. This example configures RFSA to acquire data in spectrum acquisition mode. With Reference Level: 0dBm, Start Freq: 20Hz, Stop Frequency: 3.6GHz, RBW: 3kHz and no input signal connected to the downconverter RF In port, the received spectrum shows peaks after every 50MHz bandwidth distance which is strange as these peaks don't appear when aquisition is done using NI-RFSA Soft Front Panel . This spectrum is shown in the image below.
Zoomed Spectrum..
I checked this with another example 'SMT IQ Spectrum for niRFSA.vi' from the NI Example Finder. This example configures RFSA to acquire IQ data and compute its spectrum using SMT VIs. I ran the example with the following settings; Reference Level: 0dBm, Center Frequency: 100MHz, IQ rate: 62.5MHz, RBW: 3kHz and spectrum plot from -fs/2 to fs/2. Again there was no input signal connected to the downconverter RF In port. The received spectrum again showed that peak which was appearing after every 50MHz span in the previous example plus it also showed two really high level signals near fs/2 and -fs/2. The figure below shows this
Same result was produced by the RFSA Demo Panel.vi example which is shown below
I'm not getting why are these signals appearing in the spectrum when there is no input signal connected to the downconverter? And why do they not appear on the spectrum displayed by RFSA Soft Front Panel ? Is there any configuration setting that is missing in all of these examples?
Solved! Go to Solution.
06-25-2015 04:34 AM
Try to disable the preamplifier and set the IF Filter to through using the property nodes in RFSA.
06-25-2015 07:25 AM
What you are seeing is called "Local Oscillator Leakage." The Soft Front panels use a technique called "Spectral Stitching" to mask the LO Bleedthrough. Esentially they take a span a little above and another span a litte below Fc and "Stitch" them together affer cutting out the LO. At least that's what I encountered a few years ago with the 5644R VSTs
06-25-2015 11:00 PM - edited 06-25-2015 11:01 PM
Hi Adnan,
I have checked. The Pre-amplifier is already disabled and IF Filter is by default set to Through.
06-26-2015 12:22 PM - edited 06-26-2015 12:46 PM
I think we have figured out the issue. The unwanted signals appearing on the spectrum are basically residual leakage of the dither signal. This residual leakage can be eliminated by disabling the digitizer dithering through the niscope property node. Refer to the following links
http://zone.ni.com/reference/en-XX/help/372058N-01/rfsapropref/pnirfsa_digitizerditherenabled/
Regards
09-24-2019 01:33 PM
pl ignore this email
09-24-2019 04:22 PM