Retrigger in M-Series possible ?

Hi,

since I am not able to acquire a number of sample sets with a NI PCI-6224 (M-Series) where each sample set should be started by a digital trigger : Is this card retriggerable at all ?

I tried to modify the example from the traitional DAQ : Acquire_N_-_Multi-Digital_Trig.zip (NI Developer zone) to DAQmx but did not get the code to acquire more than one sample set that starts really by the trigger itself.

Thanks,

Dirk
0 Kudos
Message 1 of 2
(2,393 Views)
Dirk,

the trigger circuitry of the M-Series boards is not retriggerable but you can use the same "trick" that you can use with the E-Series board by using the counters of the board to create a retriggerable pulse-train that can be used as the analog input scanclock.
There is an example called "Multi-Function-Ctr Retrigg Pulse Train Generation for AI Sample Clock.vi" that ships with DAQmx that does exactly what you are looking for.

Best regards,

Jochen Klier
National Instruments Germany