01-17-2007 03:15 PM
01-17-2007 03:35 PM
When there are multiple GPIB Controllers that LabVIEW can use, a prefix to address string in the form ID:address (or ID: if no address is necessary) determines the Controller that a specific function uses. If a Controller ID is not present, the functions assume Controller (or bus) 0.
LabVIEW doesn't detect any card automatically. It's just that the default for the GPIB Write (and VISA Write for that matter), is controller 0.
You might want to consider switching to VISA in any case. For quite some time, that has been the standard for which all instrument drivers are written with. With a VISA Resource Name control/constant, you can also just pick your instrument from a list. The list is populated from your Scan for Instruments in MAX.
01-19-2007 09:03 AM
01-19-2007 09:12 AM
There is no "site" for the on-line help. It's the LabVIEW help available on your pc. Go to the Help menu in LabVIEW and select Show Context Help or VI, Function & How-To Help, or on the diagram, right click on any function and select "Help".
You should have context help turned on just about all of the time when you are programming.
You should also think about switching to VISA.