PXI-5105 wrong measurement

Hello,

I'm working on a project where I have to measure a square signal of 10KHz with a max of +3V and a minus of -17V

Somehow my program doesn't do this correctly(as shown in the printscreen).

When I execute the exact same measurements with an Agligent scope, I do get the correct measurements.

 

Can someone help me with this please?

Thanks in advance

Download All
0 Kudos
Message 1 of 2
(2,808 Views)

How are you dividing the signal in order to go into the scope?  The max range of the scope is -15V to +15V.  So it will clip your negative side of the signal if you are not attenuating it.  And your configuration has a probe attenuation of 10.


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines