Hello CFER_STS
If the acquisition is finite (sample mode on the DAQmx Timing.vi
is set to Finite Samples), NI-DAQmx allocates a buffer equal in size to the value of samples per channel. So, for finite acquisition, the Samples Per Channel determines how many samples are read on each channel.
Regards
Frank R.
05-08-2015 07:39 AM
Hi CFER_STS,
I am having the same problem. Did you get it resolved yet?
SKIM2
05-09-2015 09:21 AM
07-24-2015 09:39 AM
So I made the code change to explicitly read the desired number of samples versus the contents of the buffer, but that did not change the behaviour. If NI has successfully tested a calibration using the Fluke 5700A, I would be very appreciative if you could share the code you used. I am still reading nonsense resistance values out. Also, two-wire compensation does not seem to be allowed on the calibrator when using resistances of 100 kOhm or greater, which the calibration procedure does not address. Interestingly, our in-house calibration lab attempted a calibration of the NI 9214 using NI Calibration Executive, and ran into a similar problem. Has this combination indeed been tested successfully, and if so, what code / VI was used?
Sean
10-26-2015 02:19 PM
Did you ever get any further assitance with this or figure out the error (I know its 5 months later). I am about to perform the same excercise but if it's not possible to do with the Fluke 5730A then I dont want to waste my time.
Thanks.
10-26-2015 02:44 PM