02-15-2008 10:30 AM
I have an issue configuring a Lebow load cell. I have read as much NI documentation that I can find but it is not operating properly. I have an SCC-SG24 installed in a SCC 68 box. I have configured MAX to use the SG24 Strain Gauge Module. My primary problem is that when there is no load on the Lebow Load Cell the measurement will float constantly upward and will not level off. The documentation seems to contradict itself by saying it operates in differential mode but only RSE is available. Also, I am getting a negative deflection even though I have everything wired correctly. I am using the load cell with NI’s example program Aqu Wheatstone Bridge Samples (With Calibration).vi
02-15-2008 01:52 PM
02-15-2008 02:12 PM
Matthew,
Thanks for the response! Wow, I thought I needed signal conditioning. It has 4 wires on the load cell schematic (physically on the load cell). Standard Wheatstone bridge configuration. Two for excitation and two for signal marked with both marked as +/-. So does this imply that that I just power the excitation lines with 10 volts and run the signal lines directly back to an analog I/O? That was actually working well but the singal level was very low in the order of millivolts.
02-15-2008 02:12 PM
Matthew,
Thanks for the response! Wow, I thought I needed signal conditioning. It has 4 wires on the load cell schematic (physically on the load cell). Standard Wheatstone bridge configuration. Two for excitation and two for signal marked with both marked as +/-. So does this imply that that I just power the excitation lines with 10 volts and run the signal lines directly back to an analog I/O? That was actually working well but the singal level was very low in the order of millivolts.
02-15-2008 02:34 PM
02-15-2008 02:40 PM
02-17-2008 12:20 PM
02-17-2008 05:17 PM
Logan,
Agree, I think it is the right one because it is the only one that supports 10 V excitation. Everything is attached in the original post (look at max configuration). Load cell is four wire labeled + Exc, -Exc, +Sig, and - Sig. Excitation measures 10 VDC.
02-17-2008 06:31 PM
02-17-2008 08:22 PM
Thanks! As stated I am new to load cells so tips are appreciated.
I measured the resistance and it is 350 ohms across +/- excitation and 350 ohms across +/- signal. I physically opened the connector on the load cell to confirm the wires to the pins. I am confident it is connected correctly.
The physical connections I have to the SCC SG24 are:
Exc – to SCC Pin 1 V(ex)-
Exc + to SCC Pin 2 V(ex)+
Sig - to SCC Pin 5 ACH(x)-
Sig + to SCC Pin 6 ACH(x)+
I am using a NI supplied external power to the SCC68 since I have four modules plugged in. I have verified that the SG24 V(ex) is putting out a 10 VDC voltage. My expectation was to select SCC Mod1/ai0 and I would read back the voltage at 100x gain that could be converted to a load.
I made several measurements using MAX and this is what I know.
So, what is the right mode to use with the SCC SG24?