I realize that this thread is 5 years old, but I need to state that Agilent is their wild and crazy wisedom decided to make sure their Logic Analyzers did NOT support VISA to communicate with the instruments. In fact, they made sure that VISA would not work! (insert curse here)
The instrument does not "speak" SCPI or any form of ASCII based text command; therefore it will not respond to *IDN? or any other commands from NI MAX or Agilent Connection Expert.
They use Distributed COM (DCOM) to interface with the instrument. 21st Century and there is no compatibility between instruments. AARGH!
There is a thread that discusses how to communicate with the instrument here.
02-21-2013 12:03 PM
I have a 200 page "help" file that describes COM automation with the Agilent Logic Analyzers...
02-21-2013 01:06 PM
I may need to read it 😉