10-07-2015 09:04 AM - edited 10-07-2015 09:12 AM
Hello !
I have a problem with a labview programm. The programm runs perfectly on windows xp but when it runs on windows 7 i have an error message on Keithley's display : Keithley Error Id code -222 and 822 .
It's weard because i didn't change any thing on the source code. can any body help me ?
small explication of the error message # -222 ====> Start und Stoplevel auf milivot einstellen im programm
his error occurs because there is not 10A source current capability. Example " : SOUR:CURR 10"
s # 822 parameter delta out of sens too small for sense range
-This error will be generated if you fix a measurement range and attempt to set compliance to
less than 1% of range.
Solved! Go to Solution.
10-07-2015 09:22 AM
Hi bibo,
does the "new" Win7 computer use different regional settings, compared to the older WinXP computer?
From your error explanation I guess you're using German settings in Win7. Use English settings (using point as decimal separator and comma to separate thousands)!
Or rework your Keithley driver VIs to format those numbers correctly…
10-08-2015 03:11 AM
Thanks ! i just had to change to englisch settings