What is needed to connected a Nellcor Pulse Oximeter for a adult finger to a DAQ 6008?
01-08-2013 02:25 PM
The first thing needed is for you to tell us what kind of output the instrument has. Is it an analog voltage, a logical pulse, an RS-232 port, other?
Lynn
01-08-2013 03:44 PM
The Nellcor DS-100A Adult Finger Clip Spo2 sensor has a RS-232 port
01-08-2013 04:18 PM
Then the USB-6008 is of no use. It cannot provide RS-232 communications.
You need a computer with an RS-232 port (almost impossible to find these days) or you need to get an adapter from USB or PCI (or whatever interface your comoputer has) to RS-232.
USB to RS-232 adapters using the FTDI chips seem to work well, whilc some people have had problems with the Prolific chips.
Lynn
01-08-2013 04:20 PM
You really need to verify that the sensor connection is RS-232 and not just a DB-9 style connector. Where in the manual does it say the sensor for the pulse oximeter is RS-232?
01-08-2013 05:01 PM - edited 01-08-2013 05:26 PM
I'm sorry , your are correct.
The Nellcor DS-100A Adult Finger Clip Spo2 sensor has a
DB9 7 pin connection.
01-08-2013 05:02 PM - edited 01-08-2013 05:13 PM
Is ther a finger pulse oximeter that will work with the DAQ 6008? My Projectmates and I would connect the pulse oximeter to a laptop using labview 2012 student edition to monitor heart rate and blood oxygen levels.
01-08-2013 06:41 PM
It appears that the DS-100A is just the finger sensor. The pulse oximeter instrument may be N-560 or N-600x. Both appear to have RS-232 outputs. You can certainly buy an adapter to connect RS-232 to USB for far less money than you would spend on an alternate pulse oximeter.
Lynn
01-10-2013 06:48 PM
Is there any other way to connect any finger pulse oximeter to the LabView via DAQ6008?
01-10-2013 07:06 PM