Getting error -200088 DAQmx Read (Analog DBL 1Chan 1Samp).vi<append> ...

Solved!
Go to solution

In the testpanel of NI-max we tried, but every time we are getting 10v, but the thing is if i remove the cable connection then voltage is vary in between 10 to -10v.

i have off the system power so initially i should get 0v but i didnt.

and i have connected multimeter with the pin 30 and 29 for ai3 in that i can read 0v but in control panel i see 10v every time here in attachmet i have attach that dia.

please guide me i am very confuce about this, i think i am missing some basic concept for that.

thank you very much in advance.

0 Kudos
Message 11 of 13
(1,777 Views)
Solution
Accepted by topic author AnkitRamani
In my first answer, I asked you about the configuration and you did reply. You show you are programming a diff connection but you have an rse connection wired. Change the setting in the test panel.
Message 12 of 13
(1,764 Views)

Thank you very much.

ya it might be setting problem.

I have set In test panel RSE and it seems voltage at 0v which i want and if i set differential then its show 10v  si i got confused over there.

but In RSE its show 0v which i was expect.

 

thank you very much once again.