Error -1073807339 occurred at SOSL1260 Read Single

Hi, Im working with the Solartron1254 and use LV 7.1.
I try to take a single measurement with the driver sosl 1260.
I can send datas to solartron 1254 from LabVIEW
Localy i can take a measurement with the device but if i try to read the data with LabVIEW i get this error
 

Error -1073807339 occurred at SOSL1260 Read Single

 

Possible reason(s):

 

VISA:  (Hex 0xBFFF0015) Timeout expired before operation completed.
 
This error occures at the Wait condition.vi.
 
On my device solartron 1254 i put F1=0 anfd F2=0  to Line feed.
i tried to expand the timeout but it was not working.
Can u help me?
0 Kudos
Message 1 of 4
(2,644 Views)
Are you sure that the 1254 is 100% compatible with the 1260 in terms of the command set? A timeout error could also occurr if the device doesn't understand the command that was sent by the PC.

Best regards,

Jochen Klier
National Instruments Germany
0 Kudos
Message 2 of 4
(2,640 Views)
Hello Jochen
 
 Yeah, the 1254 is not 100 % compatible to 1260 .The 1260 has some more options.
 
I found the mistake : It was the mask value was default 16 and this is wrong  if i change  it to 2 everything is ok.
 
Thanks
 
Frank
0 Kudos
Message 3 of 4
(2,633 Views)

Hi Frank,

 

I am using solartron 1255 with the drivers of 1260. I have the same error even when changing the mask from 16 to 2.

 

Any help ?

Thanks !

 

(your post is very old, I have to repost the question for a new discussion)