Hello!

I am having some issues with a Denver Instrument TP-12 analytical balance.

I have set up the balance and made sure that all the VISA settings match in LabVIEW.

 

I do manage to read accurate weight from the balance, and the read buffer works correctly, but every 30 seconds or so the read buffer gets filled with gibberish (random non-numeric characters) and so the numeric data (after converting from String to Numeric) is read as 0.

 

I contacted Denver Instrument (now Sartorius) and was told that the problem probably lies in LV.

 

Attached is the VI that I am using as well as screenshots of the read buffer when it reads the gibberish.

 

Any advice as to why I am having these weird reading and what I can do to either prevent them or a method to have the numeric data ignore the spikes in the reading would be GREAT!

 

Thanks-in-advance,

 4Strings

 

Download All
0 Kudos
Message 1 of 6
(2,756 Views)
Solution
Accepted by topic author 4Strings

Hi,

 

In order to know if LabVIEW is the culprit of this behavior, it might be very helpful to use this architecture :

other architecture.png

 

and tell us if you see the same behavior.

 

Regards,

Message 2 of 6
(2,734 Views)

It worked! 

Thank you SO much!

 

-4Strings

0 Kudos
Message 3 of 6
(2,716 Views)

Thanks for your feedback !

 

Producer/Consumer is one of the most powerful architecture for data processing/logging :

 

     - One loop (one process) to acquire datas

     - One loop for processing/logging

 

By using this architecture, you are sure to remove 99,9 % of acquisitions issues.

 

Cheers,

Message 4 of 6
(2,679 Views)
Thanks!
How would I log more data streams from other sensors (from temperature probes, condutivity probes, etc.)?
Do I just add separate ques for each of them in the producer loop, or can I add them to the first que?

Thanks again,
4Strings
0 Kudos
Message 5 of 6
(2,669 Views)

Hi 4Strings,

Mathieu_T had a great advice when he recommended the producer/consumer loop. If you want to pass multiple datatypes with the queue structure, I would recommend grouping the data into a cluster or an array, since a queue can only handle one data type (e.g. a cluster containing a string and a double). 

There's a bit more information on the producer/consumer loop here: http://www.ni.com/white-paper/3023/en/

Rick C.