01-20-2010 02:20 PM
01-20-2010 02:27 PM
Hi KowdTek
I just read about that in the manual. I will check to see if the jumpers are set correctly. Also, in the DAQ assistant where I acquire an analog signal... should I select voltage or custom voltage with excitation?
01-20-2010 02:54 PM
It's hard for me to tell because I've never used both device but it won't hurt you to try both. Whichever works for you, but I usually set just the voltage...
01-23-2010 06:09 PM
Hi
Fairly new to labview so please bare with me. I have few more questions. concerning pressure tranducer readings should I scale volts to pressure from the DAQ assistant using custom scaling or write a code in the VI? Sometimes numbers fluctuate constantly,wondering if some filtering is needed in the VI. Last question is when splitting signals from the DAQ assistant is there a way to tell which signal is what? I have temperature sensors also hooked up to my terminal blocks. Thanks!
01-23-2010 07:27 PM
@ tacky, it's fine to use scaling that can be set at DAQ Assistant, also you could tweak the samples to read and rate settings. As for how to identify which signal you are reading, use "index array" then wire the input channel right to index port.
01-23-2010 07:46 PM
If you are going to use dynamic data, then you have to be careful with the Convert From Dynamic Type. The example above assumes that there multiple channels sampled a single time. That might not be a valid assumption. That's just one reason why I avoid dynamic data. It's impossible to tell the actual data type visually
To the original question, if the split signal function is used, then the order of the outputs is the same order as the channels in the DAQ Assistant. If you are going to use something like the DAQ Assistant, then keep the dynamic data throughout and avoid the Convert From Dynamic Type unless you are very sure what you are doing.
02-04-2010 10:59 AM
02-04-2010 04:17 PM
Hi
concerning the excitation voltage needed for the cole parmer transducers, I changed to a pair of omegadyne transducers that require 10Vdc. I change jumper settings in module and Measurement&automation and think this will work better than the cole parmer. I think I can figure it out from here with all of the previous suggestions that have been given! Thanks!
10-08-2010 09:53 AM
I am using ICP pressure sensor (model 113A24) from PCB Piezotronic. I have attached the sheet. With the defined range of -5-+5V and the pressure range of 0-6895 KPa I have calculated
m = 689.5 C = 3447.5 for Y =mX+C
I have attached the sensor data sheet. Can someone recheck my calculation based on the provided data from manufacturer.
10-08-2010 10:05 AM
Those numbers appear to be reasonable. It is interesting that they specify the sensitivity as 0.725 mv/kPa, while the span calculation gives 0.6895 (although it is within the +/-0.25 mv/kpsi tolerance).
Lynn