Hello everybody,
at the moment I am working with the Agilent 34972A driver for the first time and I have a problem.
I develloped a LabVIEW VI which should log 6 temperatures and 2 voltages. The programme works fine and I receive a text data after logging. But it seems as if LabVIEW records the channels step by step and there are always about 5s between two different temperatures or voltages for each channel.
I want them to be parallelized and continuous.
It might be possible that my LabVIEW VI is the problem because I actually had no idea how to create my VI and I was really surprised when it seemed to work.
It would be great if anyone can help me fix my problem or give me some advice how to change my VI.
I attached my VI and a text data to show you how it looks like.
I want new temperatures and voltages been logged every 100ms.
I use Agilent 34972A and LabVIEW2011.
Thank you very much and if anything isn't clear yet, just ask.
09-08-2015 09:07 AM
09-08-2015 09:27 AM
Thank you for the response.
I'll try to change the program. Actually I tried to read the entire scan but I have absolutely no idea how. The problem is that I don't know how to choose and change a channel when I don't split them. Could you give me some advice?
The logging part seemed wrong for me too. I copied that from a VI I created some months ago and didn't really know how to put them together. But I'll try to work on it.
09-08-2015 09:43 AM - edited 09-08-2015 09:50 AM
You do not understand how to use that instrument correctly.
You must follow these steps.
1] Configure every channel for the measurement you want it to take.
2] Configure any scaling if needed
3] Build a scan list of the channels you want to scan
4] Scan the channels at regular intervals
The scan will read each channel sequencilay and send back the measurments
I use a state machine regularly with the 34972A
Here is an example of steps 1-3 configuring DC volts, Scaling DC shunts, K type thermocouples, and building a scan list.
To scan the instrument use the Scan-now vi that I attached. It will return a numeric array, string array, or CSV string containing each measurement.
Also don't be confused by my example using Agilent 34970 vi's the 34970 and 34972 are identical except for the communication options.
09-09-2015 06:40 AM
Thank you for the helpful response.
I changed my VI and now I can configurate the measurements.
But I still don't know how to put my 'measurement part' and my 'logging part' together.
Besides... when I schould not use the monitor.vi, which should I take instead?
09-09-2015 09:06 AM
09-09-2015 09:35 AM - edited 09-09-2015 09:36 AM
Like Dennis said look at the examples, my "Scan-now.vi" is just a modified example that came with it.
This instrument is a data logger, it is designed to take a lot of measurements at regular intervals.
It's not a fast DAQ, it is really not designed to take continuous measurements, the monitor function is there for monitoring if you need to keep a close eye on something by physically watching the instrument and that's really all.
A basic program using the 34970 or 34972 would go something like this.
Configure channels
Build scan list
Scan
Save and display measurements
Wait for next scan interval
Go to Scan
Again I find a simple state machine to be the best program architecture for this instrument.
09-11-2015 12:19 AM
Thanks again for your help.
I changed my VI and the reading and scanning works fine.
Even my data logging part works well. But both together is not working the way I want to.
I am pretty sure that I have to use a While Loop but I don't know where to put it. I tried everything put my txt data looks always like in the attached jpg.
Can you help me fix my problem? Do I really have to use a While Loop or is there a function I don't know yet?
09-11-2015 12:24 AM
09-11-2015 12:33 AM
Okay I got it.
I just deleted the two While Loops and put one around everything.
Now it works! 🙂