10-19-2012 10:15 AM
Hello All,
Does anybody know how to install an Agilent E5810 on MAX?
It works fine using Agilents IO Suite on Windows but it needs to be installed on a linux machine which Agilent do not cover. I can't find the appropriate option. The E5810 is a network device that allows a user to remotely control via IP a GPIB device.
Sean
Solved! Go to Solution.
10-22-2012 06:37 AM
Hi Sean,
Just to clarify, when you say "install Agilent E5810 on MAX" do you mean the Madrid Linux distribution (not Measurement and Automation Explorer)?
If so, that particular OS is based on either Knoppix or Ubuntu which I do not think is supported by National Instruments. Have you run any National Instruments software, or interfaced National Instruments hardware with the machine running Linux?
Kind Regards,
James O'Neill - Applications Engineer - National Instruments
10-23-2012 03:03 PM
This used to work on Windows (in NI-VISA 3.6) by adding the E5810A manually as a LAN instrument in NI-MAX using the following format:
TCPIP0::<ip address or hostname>::gpib<interface number>,<instrument GPIB address>::INSTR
For example:
TCPIP0::156.140.113.117::gpib0,22::INSTR
would talk to the E5810A gateway at IP address 156.140.113.117, the gateway is set to gpib0 (the default for the E5810A), and the instrument is at GPIB address 22. So, in NI-MAX you would add the LAN instrument at the IP address, and give it the instrument name of gpib0,22
This addressing no longer seems to work on NI-VISA 5.2 on Windows (it still works on Agilent IO Libraries Suite 16.3). This seems to be a bug in NI-VISA 5.2 for Windows.
I'm installing Scientific Linux 5.x to see if this format still works in the NI-VISA for Linux...
Alan Copeland
Agilent Connectivity Support
10-23-2012 05:46 PM
As followup to my previous post, I installed Scientific Linux 5.x inside a Virtual Box VM. I couldn't get visaconf to work (Error 13), but I was successful in using NIvisaic to talk to the instrument connected to the E5810A (using teh settings I mentioned in the previous post).
02-08-2013 11:20 AM
I am using NI-MAX v5.3.3 which doesn't allow me to use LAN Device name "gpib0,5" as it requires, instead, it only allows me to use "gpib0". Then it doesn't communicate to Agilent 6812B with GPIB address 5 via Agilent E5810. Agree with agio that it seems a bug on NI-MAX. Don't know why it hasn't been fixed.
02-08-2013 11:26 AM
Here is the screen shot.
02-11-2013 09:38 AM
Hi All,
The way to address these instruments through NI VISA has changed when VISA upgraded to version 5.2.
This KB explains how to communicate with your instruments through your LAN/GPIB gateway:
Using NI-VISA 5.2 or greater with the Agilent E5810A LAN/GPIB Gateway
Please let me know if this works for you,
James O'Neill
02-11-2013 09:42 AM
Your link won't open James, can you check it?
02-11-2013 09:47 AM
Appologies, my mistake, this one should work:
02-11-2013 03:40 PM - edited 02-11-2013 03:47 PM
I followed that article already to select"NI-VISA Passport for Tulip" and set "NI-VISA" as primary. But NI-MAX still does the same thing: don't take "gpib0,5". Please see the attached 3 screen shots.The first one captured from Agilent Connection Expert shows the correct VISA address and was recognized by Agilent Connection Expert.