Thanks for the questions.
1) is the accelerometer powered? Most accels, IEPE or ICP, require a 24V power supply with 2 to 4ma of current. If the accelerometer is not powered, then you will get a crazy FFT similar to what you describe.
2) the DC offset comes from the power used by the accel. Most vibration acquisition boards (aka - usb-9233) provide an AC coupling circuit that removes the DC. THese boards also provide the power. If you are powering externally, your DAQ board needs to have +/- 24V range to see the AC signal (vibration) riding on top of the power (DC offset = power source for accel). You can use a simple voltage divider to reduce the voltage if you need. Once you have the signal digitized, the easiet method of removing DC, is to capture enough data in one block of acquisition that you have several rotations (10 or more). You can use the statistical median then to find your DC value, and simply subtract this from your original array of data. This is less sophisticated than the high pass filter, but it works well if you have enough cycles (rotations) in your data block.
3) frequency resolution is the number of points in your FFT plot. It is also know as delta-f, or the distance in frequency of each data point in your FFT. Better resolution means smaller delta-f. You can reduce the delta-F by collecting more data (longer data acquisition time).
4) unbalance is a one times rotation vibration, it will appear in the FFT at running speed. You will need to know the speed of your machine to find the unbalance peak. Do you have a a tachometer?
Consider reviewing the following materials:
http://zone.ni.com/wv/app/doc/p/id/wv-102 |
http://www.ni.com/industrial/machine_vibration.htm
05-22-2006 07:17 AM
05-22-2006 07:31 AM
Thanks gentlemen ,
The accelerometer is not powered. It is connected to the NI 4472 that supplies the enough current to it, i guess.
i'll try the statistical median method for the DC offset.
about balancing, this is educational kit , i can connect and disconnect the weights easily to get it balanced , I also have a speed sensor , but what i am seeing here, is that the peak is not changing with the speed ( 1x rpm ) , or actually there is no specific peak! .. the FFT am getting looks really depressing! i've been studying vibration since a month and theory is far differnet than what am getting here.
Thanks one more time.
05-22-2006 07:40 AM
05-22-2006 07:40 AM
05-22-2006 08:01 AM
Thanks Lynn for the tip. Attached the file.
Thanks Preston, but i couldnt find the IEPE and coupling options , there are available at the test panels only. It seems that power is on, i mean that the signal changes when unbalanced or speed changes.
the rotor spins in the range of 1000-8000 rpm.