Hello,

I need to measure strain of 3 jauges (every jauge acquire strain of 3 directions). 

I would to know if i could use 3 unity of NI 9219 on one chassis?

Because i tried to use 2 acquisition board and i get an error! I should use only one board!

 

Thanks,

0 Kudos
Message 1 of 11
(3,373 Views)
I don't understand your question. The 9219 has 4 channels. What do you mean by 3 unity?
0 Kudos
Message 2 of 11
(3,365 Views)

Hello 

 

The picture of the unity (or module) is attached. In fact, i would to use 3 of it on the same chassis ( in the second picture). 

 

Thank you very much,

 

Download All
0 Kudos
Message 3 of 11
(3,336 Views)
What you mean to say is that you want to use 3 channels of a single 9219, if I understand your Google translation correctly. As already mentioned, your device has 4 channels so you can certainly use less than or equal to that number. For connecting them, please provide some details on the actual sensor/gauge.
0 Kudos
Message 4 of 11
(3,323 Views)

 

It is not channels! I mean 4 pieces of NI 9219 which i will use on the chassis!!

I have 9 strain gages to connect !

Thanks,

0 Kudos
Message 5 of 11
(3,314 Views)
Solution
Accepted by topic author M.A.H

Simulating a cDAQ-9174 Chassis with 3 cDAQ-9219 modules in NI MAX does not generate any errors hence I'd say yes, it is possible.

 

To make sure, I strongly recommend to contact your local NI service.

 

Regards, Jens

 

 

Kudos are welcome...
0 Kudos
Message 6 of 11
(3,303 Views)

Unfortunatly the 9219 is one of the devices that does not support a single Task with phyical channels on different devices.  You will need 3 Tasks (one for each device)

 

An Action Engine somewhat simillar to this example (where I ran into a simillar situation with multiple cDAQ devices) can help out a lot:

Flow.png


"Should be" isn't "Is" -Jay
0 Kudos
Message 7 of 11
(3,296 Views)

Thanks very much  jg69.

I would know also how would i calibrate the NI 9219.

 

0 Kudos
Message 8 of 11
(3,294 Views)
Solution
Accepted by topic author M.A.H

@M.A.H wrote:

Thanks very much  @JensG69.

I would know also how would i calibrate the NI 9219.

 


Assuming you want to account for the strain guage variabilities:  Rotate the guages and place a known mass on top.  Record the measurement with and without the mass.  Adjust the channel's "Scale" Slope and offset per actual recorded values.

 

Calibration of the 9219 is a different matter- It should have shipped with a certificate of calibration.  Well, actually, they are on-line and you look them up by serial number from "MyProducts" on your ni.com profile (you do need to register your products!)


"Should be" isn't "Is" -Jay
0 Kudos
Message 9 of 11
(3,290 Views)

Thanks 

I have registred my product now. The system must verify the product then i will download the certificate!

I would thank you another time for your answers.

0 Kudos