01-19-2016 09:37 AM
I am having trouble with gpib explorer and ENET/1000 communication.
I can happily communicate with ENET/100s, but cannot communicate with ENET/1000s. symptoms are as follows:
Searching the network for GPIB controllers does not show ENET/1000s.
I manually add the ip address (using gpib explorer and selecting enet/1000) and the message comes back that there is no such device (though I can access the configuration page through a browser).
I then tried to add manually again, but this time telling gpib explorer that the controller is an enet/100. The message comes back that this device is not an enet/100, which tells me that there is some kind of communication.
I also explicitly opened the ports listed in this http://www.ni.com/white-paper/12402/en/#toc3 recent white paper from NI (UDP port 44525 and TCP 5000, 5003, 5005, 5010, and 5015)
I am running on 64 bit Red-Hat with the latest ni4882 and firmware and both devices are on the same subnet.
All help appreciated,
Christina
01-21-2016 03:14 PM
More info:
Running on Red-hat enterprise linux 6.6
New things tried:
Taking the firwall down completely and then connecting
Installing ni4882 v 3.2 for linux instead of the most up to date 15.1
No luck
03-24-2016 08:14 AM
SOLVED: Issue addressed in realease 15.1.1 of ni4882 for linux
03-24-2016 12:15 PM
Partially addressed in realease 15.1.1 of ni4882 for linux
