genesys

sorry for the delay.

I had solved the problem a lot of time ago.

it was the cable. The genesys needs the original cable. My first cable wasn't the original one and didn't work.

0 Kudos
Message 21 of 24
(1,697 Views)

Hello, I'm trying to control the genesys TDK-Lambda 750W power supply through serial communication(Through hyperterminal). I have connected my power supply with the PC using a RS 232 cable with DB25 and Rj-45 connectors. THe power supplu has the following settings:

Bits per second:9600(Default)

Parity :None

Address:06(Default)

For serial communication the power supply is setted to use RS232 in the rear panel.

I have verified the address, the baud-rate, the switches on the rear of the instrument, but it appears all ok.

I have verified the COM in the PC, too..but it doesn't talk with the tdk lambda. when I send "ADR 06" to the power source through hyperterminal it should say "OK", but I read only

ADR 06

ADR 06

 

I don't know whether the communication port on the power supply is enabled or not.could any please help. Thank you

0 Kudos
Message 22 of 24
(1,671 Views)

Hi, the driver link you recommended uses USB as the interface, however, I do not see any USB interface on the GENH powersupply.
Does it mean I should use a RS232 to USB cable and use the driver?

Any advice is welcomed. 

0 Kudos