Why is NIMax displaying ~30 GPIB devices when I plug in my GPIB controller?

Hi everyone,

 

I have an Agilent 82357A GPIB/USB interface. That works fine and I've used it with several devices before. Now I'm trying to use it with a Keithley 236 SMU.

 

When I plug it into my computer, with NIMax open, and refresh the "devices and interfaces" tab on the left, not 1, but 30 devices appear, all of the form GPIB0::16::N::INSTR, where N ranges from 1 to 30. I've set the IEEE address on my SMU to 16, so that explains the 16 I think.

 

Here's a picture:

 

NIMAX_SMU_GPIBs.png

 

Curiously, I can communicate with the first one (it's the only one I tried so far) either using NIMax's VISA test panel so making a simple VI.

 

What's going on though, why are there 30 devices present? Should I worry?

 

thank you!

0 Kudos
Message 1 of 2
(3,923 Views)
I've not used that particular instrument but I don't think it should have that many secondary addresses. Does the manual mention any? Does the Keysight connection manager show the same thing? If you don't have another Keithley to test, I would be suspicious of the 236. It indicates a hardware fault in my experience. You should be checking with Keithley.
0 Kudos
Message 2 of 2