06-04-2013 10:03 AM - edited 06-04-2013 10:05 AM
Hello !
I'm trying to control an Ez Zone Watlow temperature controller through modbus via Labview. I've downloaded the basic VIs provided by Watlow to do this, and had them converted from v 8.6 to 8.0 (I'm running on Labview 8.2). The Ez Zone is connected to the computer (Win XP SP3) with a serial RS485 crossed cable on the PC's serial port COM1.
When I run read.vi (attached below) in Labview, and try to read a random parameter, it returns a empty string and this error :
Error 1 occurred at an unidentified location
Possible reason(s):
LabVIEW: An input parameter is invalid. For example if the input is a path, the path might contain a character not allowed by the OS such as ? or @.
=========================
NI-488: Command requires GPIB Controller to be Controller-In-Charge.
I don't really understand how this VI works, because althought it calls a .dll file, I don't see how the connection to the instrument is made.
The serial port is configured as follows :
Bits/s : 9600
Data = 8 bits
Parity : none
Stop : 1
Flow control : none
... which is the default communication configuration on the instrument.
Thanks for your help,
Paul
06-05-2013 07:48 AM - edited 06-05-2013 07:54 AM
Hi
The only problem I remember having with this controller was that the RS485 connection is half duplex NOT full duplex. I only used this to set the IP address using their supplied software as the controller had a Ethernet\ModBus adaptor and I used that for all the LabVIEW programming.
John
ps Are you running the initialize VI first?
01-26-2015 10:08 AM
I am having the same problem. Did you ever fix this?