VISA: (Hex 0xBFFF0015) occurs when I try a query to on 4400

Below is an ttachment of the screen when I try a Query to the INstron 4400.  Also attached is the NAIO trace.

 

Read Operation

Error

 

VISA: Hex(0xBFFF0015) Timeout expired before operation completed.

 

When I turn on the Instron Software to run the Instron (with IEEE enabled), it says that the program is unable to establish communication with the frame.

0 Kudos
Message 1 of 8
(5,665 Views)

Sounds like you have a problem with your instrument.  Maybe you should go call up their support.

 

But some general ideas:

Is the GPIB address in the instrument actually set to 4?

Is the GPIB enabled on the instrument?

Are you sure the instrument is connected via a GPIB cable?


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 2 of 8
(5,663 Views)

Is the GPIB address in the instrument actually set to 4? 

Yes, checked in the ADMIN of Blue Hill

 

Is the GPIB enabled on the instrument?

Yes, when "IEEE" is enabled, the machine communicates through the 488.2 card.

 

Are you sure the instrument is connected via a GPIB cable?

Yes, National Instruments 763507B-04, Type X2  GPIB.

 

I'll try their support again.

0 Kudos
Message 3 of 8
(5,660 Views)
You just said the machine communicates through the gpib card but your first post says it doesn't. Which is it? Does the instrument even support the idn query?
0 Kudos
Message 4 of 8
(5,653 Views)

Sorry for the confusion.  We recently moved the machine and when we reconnected the machine to the computer, we lost the ability to talk back and forth.  I have been trying to troubleshoot and using the VISA test panel I can get a write command to say:

 

1.  Write operation (*IDN?\n)

 

Return count: 6 bytes

 

But when It ry and read, it throws the error.

 

I don't know if this means that the write command is getting through or not.

 

I

0 Kudos
Message 5 of 8
(5,647 Views)
That seems to be an invalid command. Are you sure it's listed in the manual? Can you attach the programming manual that you're using?
0 Kudos
Message 6 of 8
(5,642 Views)

Sorry, I am not getting that deep, the VISA program makes the command.  I beleive that "IDN?" is the identification query.  Attached please find the screen that I use to try and communicate with the instron through VISA in NI MAX.

0 Kudos
Message 7 of 8
(5,637 Views)
You just can't assume the idn is supported for such an old instrument. You have to read the manual and send a command that is listed. It's not getting deep, it's getting the basics done.