10-28-2014 01:04 PM
Below is an ttachment of the screen when I try a Query to the INstron 4400. Also attached is the NAIO trace.
Read Operation
Error
VISA: Hex(0xBFFF0015) Timeout expired before operation completed.
When I turn on the Instron Software to run the Instron (with IEEE enabled), it says that the program is unable to establish communication with the frame.
10-28-2014 01:15 PM
Sounds like you have a problem with your instrument. Maybe you should go call up their support.
But some general ideas:
Is the GPIB address in the instrument actually set to 4?
Is the GPIB enabled on the instrument?
Are you sure the instrument is connected via a GPIB cable?
10-28-2014 01:25 PM
Is the GPIB address in the instrument actually set to 4?
Yes, checked in the ADMIN of Blue Hill
Is the GPIB enabled on the instrument?
Yes, when "IEEE" is enabled, the machine communicates through the 488.2 card.
Are you sure the instrument is connected via a GPIB cable?
Yes, National Instruments 763507B-04, Type X2 GPIB.
I'll try their support again.
10-28-2014 01:36 PM
10-28-2014 01:53 PM
Sorry for the confusion. We recently moved the machine and when we reconnected the machine to the computer, we lost the ability to talk back and forth. I have been trying to troubleshoot and using the VISA test panel I can get a write command to say:
1. Write operation (*IDN?\n)
Return count: 6 bytes
But when It ry and read, it throws the error.
I don't know if this means that the write command is getting through or not.
I
10-28-2014 02:05 PM
10-28-2014 02:13 PM
Sorry, I am not getting that deep, the VISA program makes the command. I beleive that "IDN?" is the identification query. Attached please find the screen that I use to try and communicate with the instron through VISA in NI MAX.
10-28-2014 02:16 PM