Trigger function in THS 710

I can't control the Trigger level from LabView. Other functions in Tk
THS7xx Config Edge Trig.vi seems to work but not the trigger level!
0 Kudos
Message 1 of 2
(2,953 Views)
Hello,
there are a few reasons why this might be happening.

- Your instrument does not have the latest firmware.
- That functionality is only supported on the 720 (the other model
supported by the driver)
- There is a problem with the sequence of commands you are using.

for troubleshooting:
- Are you getting any errors? If you haven't been using error
handling, run NI-Spy in the background to see what is going on.

href="http://zone.ni.com/devzone/conceptd.nsf/2d17d611efb58b22862567a9006ffe76/c31bc5fb34dbc0ef86256...

- Check instrument's user manual to make sure that functionality is
supported in the instrument's remote command set.
- Contact the Tektronics to make sure you are using the right series
of commands.(Lo
ok for viRead calls in your NI-Spy capture to see what
commands are sent to the 710.)