shalom,
From the specs you listed, it sounds like the PXI-2567 switch would be ideal for your application. It is a relay driver module, capable of using an external source to actuate a number of relays (up to 64) the way you described. It can accept an external voltage source of up to 50 VDC and it can provide up to 600 mA of current per channel (with a max of 25A for the whole module). Your specs (24 VDC @ 0.1A with 8 solenoids) fall well within the abilities of this board. The only possible snag is that the PXI-2567 is a PXI form factor board, so it needs to be in a PXI or cPCI chassis to function. If you cannot use a PXI chassis, you could look at the SCXI-1167, which is a very similar switch in SCXI form factor instead of PXI form factor. If you have any questions about these devices, I encourage you to get in touch with one of NI's technical representatives at 1-800-433-3488. You can post here for further discussion as well. Have a good day!
Sincerely,
RossC