SCPI-Command for PowerSupply Agilent N6700 to disable outputs if overcurrent occurs
i´m working with a PowerSupply model Agilent N6700. This is only a mainframe which can have up to 4 modules integrated.
With SCPI-commands i can enable an OverCurrentProtection for each module (1..4) what is done with this command:
CURR:PROT:STAT ON,(@4)  (where 4 is the module-number)
What i am now looking for is a command which disables ALL 4 modules immediatly  when there was anOverCurrent on one of the modules.
For example first i set all 4 modules to 10V,1A, State=ON with OCP = 1A
Then there happens during testing an overcurrent on module 2 which disables currently only the output on this module.
But i want to disable ALL outputs immediatly if there was an overcurrent on any of the modules.
I know how to do it from external by checking all states with CURR:PROT:STAT? (@1) and if one of them wozuld be true to disable manually the outputs, but i want that the mainframe is configured to execute this by itself and to disable all outputs.
Is this possible?
Thanks for your help.
Message Edited by OnlyOne on 07-26-2007 11:37 AM
(3,840 Views)
It may be faster to contact Agilent to get an answer.
(3,834 Views)
... or by reading the manual *g*:
OUTPut:PROTection:COUPle <Bool>
OUTPut:PROTection:COUPle?
Description
This command enables/disables output coupling for protection faults. When enabled, ALL output channels are disabled when a protection fault occurs on any output channel. The enabled state is On (1); the disabled state is Off (0). When disabled, only the affected output channel is disabled when a protection fault is triggered
(3,818 Views)
