Question on multiplex sampling / simultaneous sample & hold

Hello together,

 

I own a NI USB 6211 board which works very well. However I am trying to understand how exactly multiple channels are being converted from analog to digital. Therefore I drew two pictures with the way I understand multiplexing and simultaneous sample & hold. It would be great if you would check it and correct any errors I made

 

Thanks

multiplexing (?).png

 

 

and 

simultaneous sample & hold (?)

 

 

 

Can I find the exact conversion timestamp with the 6211 device?

 

Best regards



Download All
0 Kudos
Message 1 of 3
(3,482 Views)

Hello Lysandros89,

 

this could be helpfull

Setting AI Convert Clock Rate for the Longest Settling Time in DAQmx

with this properties you can read out the conversion times.

 

Additionally

Data Acquisition Sampling Terminology

 

This is an example between the difference for multiplex an simulaneous sampling how i tlooks like

Unbenannt.PNG

best regards
Alexander
Message 2 of 3
(3,464 Views)

Alright, thank you!



0 Kudos
Message 3 of 3
(3,412 Views)