Hello guillem,
Thank you for contacting National Instruments.
Open up MAX again, expand the Devices and Interfaces tree, right click on your GBIP card and select Scan for Instruments. Make sure that it successfully finds your instrument and does not give you any errors. If it is unable to find your instrument, check your cabling and connections.
If MAX finds your instrument, right click on the instrument and select Communicate with Instrument to open the NI 488.2 Communicator. In the Send String field, type "*IDN?". You do not have to precede it with IBWRT to send the command. Click the Query button. If your instrument does not send a response, check to ensure that your instrument is 488.2 compliant. If it is not, you will need to obtain the
specific command set for that instrument from the manufacturer.
If everything checks out up to this point, open up LabVIEW. Instead of using the 488.2 functions to communicate with your instrument, try using the VISA functions. I have written a small VI that illustrates how to use the VISA functions to send a command to your instrument and receive data back. I was able to locate a driver set for the MS2601 in our Instrument Driver Network. You may also want to consider trying these. They may work for your instrument.
If you are still experiencing problems, inspect your pins and cables. It is possible that one or more of your data pins may be damaged. Also, try communicating with a different instrument using the same card. Likewise, try communicating with the same instrument using a different card, if possible. All of these steps will help isolate the problem.
I hope this helps! If none of this works and you are still having problems, let me know.
Matthew C
Appli
cations Engineer
National Instruments