Trying to calibrate a PXie-5162 in Calibration Executive 3.6.4 using the Rohde&Schwarz NRP-91 Sensor. I have it so that NI Max sees it but for some reason  Cal Exec won't see it. I have the 3 software/drivers installed that is required:

         NRP Tool Kit 3.1.0

         NRP-NI-VISA Passport 2.4.1.0

         NRP-Z LabVIEW Driver 2.32

Any ideas what I am doing wrong?

Thanks in advance!

 

0 Kudos
Message 1 of 3
(3,555 Views)
Solution
Accepted by TroyI

Hi Troyl,

 

you're using the old NRP-Toolkit and the NRP-NI-VISA Passport. To work with the new NI-VISA 14+, please update your:

NRP Toolkit to 4.9

NRP-NI-VISA Passport to 2.7.0

Link: https://www.rohde-schwarz.com/software/nrpz/

 

rsnrpz LabVIEW driver to 3.4.1.0:

https://www.rohde-schwarz.com/driver/nrpz/

 

That should do the trick

 

BR

Milo

Message 2 of 3
(3,530 Views)

Worked great Thanks Milo!!!