10-07-2015 08:40 AM
Hello,
I have searched half of the internet, but I can't find any solution for my Problem. Maybe someone of you have a clue 🙂
My System Information:
OS: Windows 7 (64bit)
LabView: LabView Professional Development System 2015 (32bit)
MAX: Version 15.0.0f0
NI-VISA: 15.0
NI-Serial 14.0
I have attached a screenshot of MAX at the end of my post.
All windows drivers for the serial ports are correctly installed (see other screenshot). But in MAX, I can't find these ports. There is only the error code "0xBFFF0011 (VI_ERROR_RSRC_NFOUND)". If I want to add the COM Port via "Devices and Interfaces" --> "New" --> "Port (Serial or Parallel)", I can only choose between COM1-COM4 or LPT1-LPT4. But I need COM9!
Does anyone have any idea what I do wrong?
Thank you
deZeus
Solved! Go to Solution.
10-07-2015 09:57 AM
@deZeus wrote:
Hello,
I have searched half of the internet, but I can't find any solution for my Problem. Maybe someone of you have a clue 🙂
My System Information:
OS: Windows 7 (64bit)
LabView: LabView Professional Development System 2015 (32bit)
MAX: Version 15.0.0f0
NI-VISA: 15.0
NI-Serial 14.0
I have attached a screenshot of MAX at the end of my post.
All windows drivers for the serial ports are correctly installed (see other screenshot). But in MAX, I can't find these ports. There is only the error code "0xBFFF0011 (VI_ERROR_RSRC_NFOUND)". If I want to add the COM Port via "Devices and Interfaces" --> "New" --> "Port (Serial or Parallel)", I can only choose between COM1-COM4 or LPT1-LPT4. But I need COM9!
Does anyone have any idea what I do wrong?
Thank you
deZeus
If Windows does not see your device, then NI-MAX certainly won't.
10-07-2015 10:11 AM
@nyc: Yeah. You are right. If Windows does not see my device, then NI-MAX certainly won't.
But as you can see a few lines above, all necessary drivers are correctly installed and Windows "see" my devices (as you can see in the attached screenshot). Only MAX can't see any serial ports.
10-07-2015 10:50 AM
I looks like something is wrong with your NI VISA install (since MAX isn't finding ANY ports). Make sure NI VISA is enabled (sometimes other kinds of VISA get in the way) or try reinstalling NI VISA. The NI Serial is not needed since you are not using NI serial ports.
10-07-2015 11:14 AM - edited 10-07-2015 11:16 AM
Thank you for your fast response.
NI VISA is enabled (so far as I can see). I have installed NI VISA 24h ago, but I'll do it tomorrow again. I hope it will resolve the problem.
deZeus
10-08-2015 06:25 AM
Thank you crossrulz!
First I deinstalled:
After that, I have re-installed only NI SignalExpress. And now, as you can see in the attached sreenshot, MAX shows me all serial ports!
Thank you!
deZeus