08-15-2014 02:49 PM
There was a previous post about issues of the Agilent 82357A in Windows 7 http://forums.ni.com/t5/Instrument-Control-GPIB-Serial/MAXVISATimeoutError-when-82357A-USB-to-IEEE48...
Ours is different in that while the adapter is detected in both NI-MAX and Agilent Connection Expert and we are able to communicate with an instrument using them,
LabVIEW 2013 64-bit running in Windows 8.1 does not see any instrument that is connected. In other other words, when a VISA resource is dropped on the front panel, the instrument is not displayed.
Any ideas on the cause?
08-15-2014 04:04 PM
08-15-2014 07:36 PM
@Dennis_Knutson wrote:
Have you tried with 32-bit LabVIEW? Curious whether it's isolated to the 64-bit version.
Not yet.
Late in the day, I installed the 32-bit version. Will try it out on Monday.
08-19-2014 08:27 AM
Update:
The Agilent 82357A connected to an instrument is recognized in LabVIEW 2013 (32-bit).
Would welcome feedback on how to get this USB-GPIB adapter working with LabVIEW 2013 (64-bit).
08-20-2014 08:46 AM
There is no Tulip passport for NI-VISA 64bit so NI-VISA 64bit won't recognize any Agilent GPIB controller.
The deal with 64-bit VISA is similar to COM-VISA. There is a Global Manager and multiple (any company that wants to supply a 64bit VISA) vendor specific resouce managers.
I think you have to install Agilent VISA to get this to work. I know this all worked in 32bit land but that was because of the TULIP passport.
08-20-2014 09:23 AM
For now we will be sticking with NI-VISA and use LabVIEW 2013 32-bit, but we may be forced to revisit this in the future.